HEPA filter leak test


(Manoj Singh) #1

PAO has 0.1 to 1.0 micron particle size.which range of particle are scanned by scanner to detect the leakage in HEPA filter.


(vaibhav dongare) #2

scanning is related to concentration used. its not related to particle size. limit - downstream leakage should not exceed 0.01% concentration


(Nguyen hoai giang vu) #3

The scanner measuring base on the concentration of the smoke. But the smoke generator create the smoke with the size from 0.2-0.3 micro. And the hepa filter have the efficiency 99.97%at particle size of 0.3 micro.